Requests

support
Showing: limited to 100 requests
Modification filters
(2 responses) Zimeng 2024-11-26 23:40

Hello Skyline team.
I am working on post-translational modifications and am using skyline to export PRM lists. I would like to generate a list of peptides containing only a specific modification based on the peptide search results, is there a way to do this?
Unfortunately, search results exported from PD cannot be filtered for specific PTMs and .pdResult files imported into Skyline cannot be filtered for specific PTMs.

Yours,
Zimeng

view request
Multiple spectral libraries within the same Skyline doc
(3 responses) lincolnh 2024-12-05 10:51

Hi Skyline team,
I have a use case where I'm comparing the performance of several spectral libraries for the same set of raw files. It would be great if I could compare different spectral libraries within the same Skyline document without reimporting the raw files. That is, I want to keep the settings in the doc exactly the same and just compare the results of loading spectral library A vs spectral library B.

I know that I can go to Peptide Settings > Library > Edit to import a new library, but I get the feeling that Skyline doesn't actually update the settings according to the new library. The only way that I've been able to successfully perform this task so far is by creating separate Skyline docs for each library, which is cumbersome and slow.

Is there anyway to do this within the same Skyline doc?

Thanks,
Lincoln

view request
Wiff2 files of MS/MS data from SciexOS
jeroen kooistra 2024-12-06 01:30

Dear Skyline team,

Two years ago a former colleague of mine posted a question titled: "Sciex TripleQuad7500 .Wiff2 support" (roosso 2022-07-05 22:41).
We still encounter this problem, also with the newest Skyline versions.
Some methods only generate ".wiff2" files. We think this happens when we apply experiment scheduling (3.5min of ESI- followed by 3.5 min ESI+)
On import of the .wiff2 file in Skyline we get the error below:

"Failed importing results file ...
The sample .... contains no usable data."

This also happens for methods where both the .wiff and .wiff2 files are generated and we try to import the wiff2 file.
I attached a Skyline file which we use for our 'System Suitability Test', including some raw data we would like to import. The .wiff file succeeds, the wiff2 file fails to import.

The last two years we used work arounds, so that SciexOS would generate .wiff files.
At moment we develop a method which requires experiment scheduling, and so will only generate .wiff2 files.
Could you please advise how to import wiff2 files?

Thanks a lot!

Regards,

Jeroen Kooistra

 SST data for test.zip 
view request
An error occurred attempting to export. ERROR: Registry key (Software\Thermo Instruments\TNG\Stellar) not found. Stellar is not installed on this machine.
(2 responses) skyfall 2024-12-05 03:49

Hi,
Using the neat feature of the export (multiple) methods. Since Skyline is usually installed on personal computers and not on the MS-linked computer, it would be great to generate those methods anyway. Is this a thermo requirement, or is there a workaround to install the necessary software on the user PCs as well?

thank you!

Skyline-daily

An error occurred attempting to export.
ERROR: Registry key (Software\Thermo Instruments\TNG\Stellar) not found. Stellar is not installed on this machine.

Command-line: Method\Thermo\BuildThermoMethod -t -m "W:\User\yx.meth" "..transitions.txt"
Working directory: .....\

OK More Info

Skyline-daily (64-bit) 24.1.1.284 (bc93c2813)

System.IO.IOException: ERROR: Registry key (Software\Thermo Instruments\TNG\Stellar) not found. Stellar is not installed on this machine.

view request
Chromatogram information unavailable
(3 responses) yaoyao wang 2024-12-04 18:15

About 8 peptides were tested by waters QTof under MSMS mode. Peptide 1 resulted in "Chromatogram information unavailable" When I used skyline to analyze raw data, while others could work under same settings.

I am confused and want to know how to solve this.

I uploaded two peptieds information and results(peptide 1 and 2) in the file.

 20241204 pep for skyline questions.sky 
view request
skyline full documentation file for Claude/ChatGPT/AI knowledge-base
(4 responses) skyfall 2024-12-02 02:06

While you provide great tutorials and fantastic support, I was wondering if a readme is available with the full documentation of all features, settings, GUI usage, etc.? That would be fantastic for knowledge-based chatbots. Especially in the beginning, where finding some of the features sometimes can be a a little difficult.

Tanks you,
best
André

view request
How to calculate signal/noise ratio for LC peaks?
(2 responses) rangika gurunnanselage 2024-12-03 16:16
I am using Skyline for lipidomic analysis and would like to set the limit of detection at 10, based on the signal-to-noise ratio. Could you explain how to calculate the signal-to-noise ratio for a specific peak in a chromatogram?
view request
Skyline failed writing error when saving .sky files
(1 response) monteb 2024-12-03 11:48

Getting error message when saving Skyline .sky files to our file server.

Error states "Failed writing to ... the process cannot access the file because it is being used by another process".

Please advise

 Skyline failed writing issue.3.png 
view request
Isotope Corrections
(1 response) annikasilverberg 2024-12-03 09:23

Hello,

I was in the lipidomics webinar today and was curious if there's a way to do isotope corrections and isotope abundance calculations directly in Skyline?

Thanks!

view request
Untargeted lipidomics using Skyline
(1 response) nidia lauzon2 2024-12-03 09:17

Hi, I would like to use Skyline for untargeted lipidomics analysis. I have an Exploris240, and LipidSearch for lipids identification. I am doing AcquireX for high-throuput lipidomics. I am also using more than one internal standards for lipids class normalization.

I have never used Skyline, and I would like to know if you have a Tutorial for that.

Thank you very much
Nidia

view request
Error in importing raw results due to modified .skyd file
(1 response) hana selicova 2024-12-03 11:02

Good evening,
lately I am unable to import raw files to the Skyline (I just uploaded the newest version). Everytime I try to upload my result files by the usual way, it starts importing and and after few seconds each of the files gives the "failed" which imideately turns into the "canceled" yellow saying. Sometimes first few injections upload, but the rest is always blocked from importing. Everytime the data are not open anywhere else. I tried rebuild the .skyd cache, reexporting the files in case there would be problem in them, saved the Skyline for those data in different folder in case there would be the clash there, checked the read/write permissions. But here I am done with what I had in my mind that could possibly be the problem. Could you help me solve this issue please? I am attaching the error statement.

 skyline error.png 
view request
SureQuant Method development
(5 responses) darora 2024-12-02 03:50

So I´m facing some method development issues for Surequant with heavy labelled peptides. I followed the tutotrial for Surequant but when analysing my files, I saw there was no signal for my endogenous peptides which were spikedin at 250 ng. Could there be a specific reason for this?

view request
Bruker timsTOF HT ion mobility filtering resolving power
(1 response) chloe baldreki 2024-11-28 08:17

I'm very new to Skyline and am trying to set up my project for some DIA data collected on a Bruker timsTOF HT. In doing so I have a few questions...

  1. When configuring the full-scan settings, is it best to use the resolving power that is stated by Bruker for this instrument (60,000)? I ask as I read in another post that the resolution setting in Skyline is used to define the width of the XIC extraction, and if in MS1 it is set a little lower than the instruments stated resolution, it can account a bit better for potential mass accuracy deviation. If going a little lower is useful, do you have any suggestions on what this should be?
  2. How do I determine what the ion mobility filtering resolving power should be? Is this something I should be able to ascertain from my data?
    Thank you for your help!
view request
Import results from FRAGPIPE to create spectrl libraries
(10 responses) afshari1 2024-10-01 12:47

Hello,
I am wondering if we can import results from FRAGPIPE to create spectral libraries on Skyline. If yes, what formats are supported by Skyline.

Thanks

view request
GC MS Agilent import
(5 responses) daria makeeva 2024-11-25 07:48

Hello,

I am trying to set up Skyline for the analysis of GC-MS data from an Agilent single quadrupole instrument. I followed Pawel Sadowski's protocol for Shimadzu and also tried adapting the guidance provided here: https://skyline.ms/announcements/home/support/thread.view?rowId=43600. However, no matter what changes I make to the transition list, the files do not process correctly. Skyline identifies only the precursor ion and cannot detect the fragments.

That said, when I manually click on the chromatogram, I can see that the MS/MS spectra have been recorded and are readable from the files.

I have attached the Skyline file, the GC-MS data file, and the transition list to make it easier to review the data.

Thank you very much for your help.

Best regards,
Daria

 GC-MS_Skyline.zip 
view request
how to detect peaks with the same MRM transition
(17 responses) a van dam 2017-01-24 01:30
For our metabolomics project we always made use of Multiquant.
My colleagues, working with proteomics, are very content about skyline for quantification.
I wanted to compare the multiquant output with the skyline output for my project, but there seems to be a problem.
The acquisition method contains some identical MRM transitions for different metabolites such as for ex D-ribose-5-phosphate and Ribulose-5-phosphate. We can distinguish because of the difference in RT.
In Skyline the results for Ribulose-5-P at RT 7.2 min seems to be okay, but there are no data for D-rib-5-P at RT 6.3 min.
Even that part of the chrom is not present, it starts at 6.6 min instead of 6.0 min
In Multiquant both of the components have been quantified.

We can probably overcome this problem to adjust the masses of the MRM transitions in the acq method, but then also Multiquant has to be adapted.

Another question is the way of integration. Sometimes it would be preferable to draw an angled baseline instead of straight basline at the x-axis. Is there a tool to do that?

With kind regards,
Annie van Dam
 170119vgl Skyline en Multiquant.pptx 
view request
MaxQuant error to build a library
(1 response) melo-braga 2024-11-26 01:30

I am trying to create the library with msms.txt file from MaxQuant but I keep receiving a error. What do I need to do in order to proceed?
Thank you!

 Screenshot.jpg 
view request
Is it possible to optimise % HCD and % CID collision energies
(3 responses) alan atkins 2024-11-20 12:25

Similar to the existing routine for optimizing CE on a triple quad, is it possible to optimise % HCD and/or % CID on a Thermo Stellar MS?

view request
No calibration curve
(7 responses) akhilabrai 2024-11-17 03:13

Dear Skyline team,

I am currently working on generating calibration curves for several peptides. Despite observing good peak areas, I am unable to obtain a calibration curve for some peptides. I have attached a screenshot for your reference.

Additionally, I would appreciate guidance on selecting the appropriate quantification settings, as there are multiple options available (e.g., linear, bilinear). Could you please advise on which method is recommended and how to approach LOD calculation? Is it possible to choose different quantification options for individual peptides?

Thank you.
Akhila

 Calibration curve.jpg 
view request
Peaks being Cut-Off on Skyline
(1 response) achiu1 2024-11-22 13:27

Hello,

I'm running a DDA skyline file to look at certain peptides and it's giving me cut-off peaks near the middle. I'm using an orbitrap mass analyzer to run my samples. The peak and transition settings are attached below. The same samples were ran with PRM and the peaks looked completely fine.

 Capture1.JPG  Capture2.JPG 
view request
Peptide Uniqueness not being enforced
(4 responses) pliuni1 2024-11-21 09:23

Hi Skyline Team,

I am having an issue in my document regarding the "Enforce Peptide Uniqueness by: Proteins" function.

My target list contains a list of peptides that I thought were unique, but when I copied the peptides and pasted them back in the document, some were flagged as also being present in the background proteome. The proteins I am building a targeted method with are recombinantly expressed and are from another organism entirely.

It was my understanding that when you enforce peptide uniqueness by protein, any peptides in the background proteome should be removed from the target list of peptides, but this does not seem to be happening in my case.

When I select Edit>Unique Peptides, I can see that some of the peptides are shared with background proteins. So I am unsure as to why they are not being removed when I enforce peptide uniqueness.

Any help on this would be greatly appreciated.

-Peter

view request
GPF_runs: No matches passed score filter in
(7 responses) vanyabangera 2024-11-19 21:52

Hello Skyline team,

I am trying to build a chromatogram library using GPF runs by setting the parameter "run peptide search." When I finish, I get the "No matched passed score filter in." How do I solve this issue?

Thank you
Vanya

 No_matches.PNG 
view request
Can different scan filters (e.g. HCD or CID) of the same precursor be separated?
(1 response) alan atkins 2024-11-20 12:31

If data has been acquired using an acquisition method that uses both HCD (beam type fragmentation in a collision cell) and CID (Resonant fragmentation in an Ion Trap) on the same precursor, can those filters be separated in Skyline? Currently, when I try to process with Skyline, it treats the MS2 fragments the same, so the you get a noisy combined XIC.

view request
Error when Merging Skyline Documents with Same Target lists, when using iRT molecules
(5 responses) Will Thompson 2024-11-19 14:12
Hi Skyline Team,

I've got what I think may be a buggy edge case where I can't seem to be able to merge two skyline documents. The documents are attached. The error I get is pasted below, and I think it might be because I have some of the compounds listed twice in each Skyline document, once to use them for iRT alignment, and the second time as an internal standard. (It's a bit of a problem in Skyline currently that you don't allow a molecule to be used both as an iRT marker, and an internal standard for quantitative purposes...therefore I need to list some molecules twice under different molecule groups).

Thanks for looking into this. If it's a bug you can help me work around in the interim by 'force' merging these two documents for me, I'd be indebted.

Cheers

Will

Skyline-daily (64-bit) 24.1.1.284 (bc93c2813)

System.IO.InvalidDataException: The peptide 13C4/15N2 L-Asparagine (Asn ) was found multiple times with user modifications. ---> System.IO.InvalidDataException: The peptide 13C4/15N2 L-Asparagine (Asn ) was found multiple times with user modifications.
   at pwiz.Skyline.Model.SrmDocument.MergeMatchingPeptidesUserInfo(IList`1 peptideGroupsNew) in C:\proj\pwiz\pwiz_tools\Skyline\Model\SrmDocument.cs:line 1353
   at pwiz.Skyline.Model.SrmDocument.ImportDocumentXml(TextReader reader, String filePath, MergeAction resultsAction, Boolean mergePeptides, FindLibrary findLibrary, MappedList`2 staticMods, MappedList`2 heavyMods, IdentityPath to, IdentityPath& firstAdded, IdentityPath& nextAdd, Boolean pasteToPeptideList) in C:\proj\pwiz\pwiz_tools\Skyline\Model\SrmDocument.cs:line 1262
   at pwiz.Skyline.SkylineWindow.ImportFiles(SrmDocument docOrig, ILongWaitBroker longWaitBroker, IList`1 filePaths, MergeAction resultsAction, Boolean mergePeptides, IdentityPath to, IdentityPath& firstAdded) in C:\proj\pwiz\pwiz_tools\Skyline\SkylineFiles.cs:line 2671
   at pwiz.Skyline.SkylineWindow.<>c__DisplayClass833_0.<ImportFiles>b__2(ILongWaitBroker longWaitBroker) in C:\proj\pwiz\pwiz_tools\Skyline\SkylineFiles.cs:line 2621
   at pwiz.Skyline.Controls.LongWaitDlg.RunWork(Action`1 performWork) in C:\proj\pwiz\pwiz_tools\Skyline\Controls\LongWaitDlg.cs:line 254
   --- End of inner exception stack trace ---
   at pwiz.Skyline.Util.Helpers.WrapAndThrowException(Exception x) in C:\proj\pwiz\pwiz_tools\Skyline\Util\Util.cs:line 1920
   at pwiz.Skyline.Controls.LongWaitDlg.PerformWork(Control parent, Int32 delayMillis, Action`1 performWork) in C:\proj\pwiz\pwiz_tools\Skyline\Controls\LongWaitDlg.cs:line 202
   at pwiz.Skyline.SkylineWindow.ImportFiles(String[] filePaths) in C:\proj\pwiz\pwiz_tools\Skyline\SkylineFiles.cs:line 2629
   at pwiz.Skyline.SkylineWindow.importDocumentMenuItem_Click(Object sender, EventArgs e) in C:\proj\pwiz\pwiz_tools\Skyline\SkylineFiles.cs:line 2579
view request
Issue, with skyline candidate peaks.
(3 responses) anirudhkashyap511 2024-11-19 10:28

Subject: Assistance with mProphet Peak Picking in Skyline

Dear Skyline Team,

I hope this message finds you well.

I have encountered some instances where Skyline appears to overlook clear candidate peaks, which results in the mProphet peak picking model selecting incorrect peaks. In the attached screenshot, you can see an example of a very straightforward peak that Skyline does not even recognize as a candidate.

Could you please advise on possible reasons for this behavior and suggest any steps I could take to resolve it?

Thank you for your assistance!

Best regards,
Anirudh

 wrong_peak.png 
view request
Waters Xevo G2 QTof raw file error
(4 responses) bruce37661 2024-11-17 18:17

Hi, I am new to Skiline and beleive I must be missing something when importing my raw file. I can open the file in MassLynx with a problem, however, the file looks like it is importing up until the very end and then gives the error message.

At 6:07 PM:
Failed importing results file 'C:\Users\RobertRegister\Dropbox\Skyline\Tocilizumab 61 min scan.raw'.
error reading spectrum function=1 process=0 scan=24757
Under More Info it says

At 6:07 PM:
Failed importing results file 'C:\Users\RobertRegister\Dropbox\Skyline\Tocilizumab 61 min scan.raw'.
error reading spectrum function=1 process=0 scan=24757
pwiz.Skyline.Model.Results.ChromCacheBuildException: Failed importing results file 'C:\Users\RobertRegister\Dropbox\Skyline\Tocilizumab 61 min scan.raw'.
error reading spectrum function=1 process=0 scan=24757 ---> System.Reflection.TargetInvocationException: error reading spectrum function=1 process=0 scan=24757 ---> System.Exception: error reading spectrum function=1 process=0 scan=24757 ---> System.Exception: [pwiz::CLI::msdata::SpectrumList::spectrum] Unhandled exception: EOF Scan File Read Error
at pwiz.CLI.msdata.SpectrumList.spectrum(Int32 index, DetailLevel detailLevel)
at pwiz.ProteowizardWrapper.MsDataFileImpl.GetCachedSpectrum(Int32 scanIndex, DetailLevel detailLevel) in C:\proj\skyline_24_1\pwiz_tools\Shared\ProteowizardWrapper\MsDataFileImpl.cs:line 1384
at pwiz.ProteowizardWrapper.MsDataFileImpl.GetSpectrum(Int32 spectrumIndex) in C:\proj\skyline_24_1\pwiz_tools\Shared\ProteowizardWrapper\MsDataFileImpl.cs:line 995
at pwiz.Skyline.Model.Results.SpectraChromDataProvider.Spectra.ReadSpectrum(Int32& i) in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\SpectraChromDataProvider.cs:line 1109
--- End of inner exception stack trace ---
at pwiz.Skyline.Model.Results.SpectraChromDataProvider.Spectra.ReadSpectrum(Int32& i) in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\SpectraChromDataProvider.cs:line 1160
at pwiz.Skyline.Model.Results.SpectraChromDataProvider.Spectra.Read() in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\SpectraChromDataProvider.cs:line 987
at pwiz.Skyline.Model.Results.SpectraChromDataProvider.Spectra.<RunAsync>b__33_0() in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\SpectraChromDataProvider.cs:line 845
--- End of inner exception stack trace ---
at pwiz.Skyline.Util.Helpers.WrapAndThrowException(Exception x) in C:\proj\skyline_24_1\pwiz_tools\Skyline\Util\Util.cs:line 1926
at pwiz.Skyline.Model.Results.SpectraChromDataProvider.Spectra.NextSpectrum() in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\SpectraChromDataProvider.cs:line 950
at pwiz.Skyline.Model.Results.SpectraChromDataProvider.ExtractChromatogramsLocked() in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\SpectraChromDataProvider.cs:line 273
at pwiz.Skyline.Model.Results.SpectraChromDataProvider.ExtractChromatograms() in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\SpectraChromDataProvider.cs:line 253
at pwiz.Skyline.Model.Results.SpectraChromDataProvider.SetRequestOrder(IList`1 chromatogramRequestOrder) in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\SpectraChromDataProvider.cs:line 617
at pwiz.Skyline.Model.Results.ChromCacheBuilder.Read(ChromDataProvider provider) in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\ChromCacheBuilder.cs:line 388
at pwiz.Skyline.Model.Results.ChromCacheBuilder.BuildCache() in C:\proj\skyline_24_1\pwiz_tools\Skyline\Model\Results\ChromCacheBuilder.cs:line 256
--- End of inner exception stack trace ---
Any idea what is going on? Also, it says Chomatogram information unavailable.

Any help would be appreciated !! Bruce

view request
Using Surrogate internal standard
(4 responses) jihyunk 2024-10-31 14:14

Hi,

I'm currently quantifying small molecules using Skyline. For some analytes, I have labeled internal standards, but for others, I don't. So, I'm considering using surrogate internal standards from different molecules. I'd like to assign these individually but haven't found instructions on how to do this.

Could you please guide me on where to find this information?

Thank you.

view request
Set non-specific digestion for the Background proteome
lxiiaanog 2024-11-19 03:00

Dear Skyline,

Is there a way to set digestion enzyme as non-specific for the Background proteome? In the context of peptidome analysis having a non-specific background would be quite useful.

Thanks in advance,
Li

view request
Importing DIA-NN spectral libraries into Skyline
(6 responses) yuxing xue 2024-11-07 02:19

Dear Skyline,

I am now encounting an issue when I am importing the spectral library into the Skyline.

I used the .speclib format file and report.tsv files that are generated from the DIANN and renamed them as "NANE.speclib" and "NAME_report.tsv". I imported the spectral libraries through File-Import-Peptide Search-Add files...Then it came an error information which asked me if that report.tsv is the correct one for the .speclib file. Do you know what might be the potential cause for this issue and how I can slove it?

FYI I am using DIA-NN 1.8.2 beta 34 and Skyline daily versions. The error information is also attached.

Thanks in advance!

Best wishes,
Yuxing

 ERROR.png 
view request
LOD and LOQ calculation
(1 response) a das 2024-11-16 04:49

Hi Nick,

I am doing ToF MRM for peptide quantitation in my sample with heavy-isotope labeled peptides.

  1. Can you please help to calculate LOD and LOQ in skyline?
  2. CV calculation?

Please share any tutorials having all these

Thanks
AD

view request
initiator methionine peptide
(4 responses) dkueltz 2016-01-11 01:34
I am trying to use N-terminal peptides for quantitation. The initiator met is cleaved off cotranslationally for many proteins and, as a result, the peptide is not properly recognized as a tryptic peptide by Skyline (since it is not preceded by a K or R but by M) if we digest the proteome with trypsin [KR | P]. When digesting the proteome with trypsin-CNBr [KRM | P] then cleavage also occurs at internal M and the corresponding peptides are incorrectly marked as having a missed cleavage. Is there any way to define a digestion scheme that is based on trypsin [KR | P] but also includes cleavage of the initiator methionine, which is common for many proteins? My lab is working on quantifying the N-terminal peptide because we study effects of enivronmental stress on N-terminal modification (acetylation, etc) of proteins.
Happy 2016!
Dietmar
view request
Creating & naming spectral libraries from ssl files via command line
(1 response) Jonas Becker 2024-11-13 10:18

Dear Skyline Team and Community,

I'm generating spectral libraries from ssl files via the command line tool with the following (simplified) code:

"D:\Software\Skyline\SkylineRunner.exe" ^
--in="tool_comparison.sky" ^
--import-search-file="results_tool1.ssl" ^
--out="tool_comparison.sky

How can I set the name of the generated spectral library? Currently, it is automatically named "tool_comparison.blib". Is there any option to assign a name or have it named after the ssl file (i.e. "results_tool1.blib")?

Additionally I receive the following warning: "Warning: Unable to locate results file 'results_tool1.mgf'". This does not seem to cause any major error for my analysis, nevertheless it would be nice to understand why this is happening. The mgf file is in the same folder as the ssl file and naming in the ssl file is correct.

Thanks in advance! Cheers
Jonas

view request
Importing lists of peptides using SkylineRunner.exe / CMD
(2 responses) Jonas Becker 2024-09-09 08:25

Dear Sykline-Team and Community,

I'm looking for help in streamlining some skyline analysis using the SkylineRunner.exe / command line version.

I have a list of peptides in txt format (peptides.txt file attached). Is there any possibility to add these peptides to the Skyline document using SkylineRunner.exe? When analyzing the data using the GUI, I just copy them into the target list which creates a "peptides1" entry with all of the peptides - something similar would be needed.

I do have a spectral library for all the peptides in "peptides.txt" which I add using --add-library-path, peak boundaries from a search engine which I add using --import-peak-boundaries and a custom report which I can also create using --report-add. The only thing missing to have this completely in CMD is adding the peptides to the document.

Additionally, I add the corresponding raw files using --import-file and --import-replicate-name, i.e.

"SkylineRunner.exe" ^
--in="test.sky" ^
--import-file="condition_untreated1_20220522.raw" ^
--import-replicate-name="untreated" ^
--import-file="condition_treated1_20220522.raw" ^
--import-replicate-name="+treatment_1" ^
--out="id_all.sky"

However when doing this, the two rawfile end up both in the replicate "+treatment_1". How to have them in separate replicates as defined in the command? When I run the imports after each other, it works fine.

Thanks in advance for you help!
Jonas

 peptides.txt 
view request
Assigning peak colors for small molescules
(5 responses) sarah williams 2020-08-06 15:50

Hello,

I am working on a project with tmt labeled small molecules. I would like to assign colors for the MS1 peaks I have assigned for each of the small molecules and their tag. I noticed there is a tutorial for this for peptides, but I can not seem to make it work for my small molecules.

 Colors Peaks skyline support.JPG 
view request
Analyte concentration to heavy
(4 responses) a das 2024-11-11 20:54

Dear Skyline support,

I am looking for peptide quantitation in my sample using heavy standards of my peptides. After making the calibration curve based on serial dilutions of heavy standards, if selected normalization method as heavy in the quantification tab, I am not getting R2 of 0.9. While keeping quantification method as none, I am getting R2=0.9.

  1. Am I doing something wrong in making calibration curve?
  2. I am not able to see the peptide quantitation in (nmol) normalized to heavy.
  3. In the calibration curve, peak area is there on y axis but not light to heavy peak area ratio.

Please guide me, which settings I need to follow to get my results.

Thank you
Arpita

 r2 not 0.9.tif  r2=0.9.tif 
view request
New DIA-PASEF method process with DIANN and export to skyline
(7 responses) JeromeVia 2024-10-14 06:05

Since diann 1.9 version, diann results could be directly export to skyline. Thanks to this very useful option.
To the slice-PASEF or Diagonal-PASEF acquisition, skyline is able to extract MS1 scan of the peptide but fail to extract MS2 ions.
Something was wrong in my parameters or skyline is not compatible to these acquisition methods?

Thanks
Jerome

view request
Issue with the spectral library building
(3 responses) Anjana 2024-11-06 02:51

Dear Skyline team,

We were trying to build a spectral library using DIA raw files in Skyline v24.1 through import->peptide search. In the build option, while we were trying to add the raw files, it seems the file format was not supported. In the previous versions of the Skyline, we have found an option to "start from -> DIA raw files". Kindly guide us to generate a spectral library using the DIA raw files in the recent version.

 Error.PNG 
view request
Batch creation of SCIEX methods
(2 responses) kerteszv 2024-11-05 13:45

SCIEX TOF

I have a set of precursor masses and/or formulas (say n different one), and a set of collision energies (CE) (say m different one). I would like to create n x m individual method files, all with different precursors and CEs. The name should somehow represent the formula/precursor mass and CE, say sample_precursormass_CE.dam

I see that maybe it might be done in a batch using
BuildAnalystFullScanMethod.exe
but I cannot figure out in what format I need to provide the precursor/CE list, and what parameters I have to use with the EXE. I know that I need a file with the precursor and CE values, and a template method file. I was able to do one inside the software, but even if the CSV file has multiple lines, it just created new method for the first line. If you have a tutorial on that or a video, I would appreciate that, too, I could not find it.

If you could send me an example command line , that would be appreciated.

Thank you very much in advance, Vilmos

PS: Bonus: if I could also do the same, PRM methods with given inclusion lists for a Thermo QExactive, that would be phenomenal.

view request
Koina error
(1 response) ericw04 2024-09-27 12:01

Hi,

I'm trying to run an EncyclopeDIA search through Skyline with an experiment-specific FASTA file. This has worked for me in the past, however, when I run the search now I get a "GO AWAY" error from Koina before it's finished processing the FASTA and the search fails. The text of error is attached. When I check the Koina configuration in Skyline (Skyline (64-bit) 24.1.0.199 (6a0775ef83)) it says the connection is active.

Any help is much appreciated.
Thanks,
-Eric

 Koina_go away.txt 
view request
Skyline LOQ = Blank plus 10*SD
karamj33405 2024-11-04 10:38

We were curious if it would be possible to add a different option for Skyline's Molecule Quantification LOQ . The Max LOQ CV is useful, but typically for our PFAS quantitation it is more common to see the LOQ calculated like the LOD, which would calculate LOQ based on Blank plus 10 * SD.

  • Kara Joseph, Baker Lab at UNC Chapel Hill.
view request
MS² Peak Intensity Prediction for phosphopeptides
(6 responses) afshari1 2024-11-02 14:25

Hello,
I imported an MSP format spectral library on Skyline as the Human Phosphopeptide Label Free Library. However, it does not show any spectra for the phosphopeptide target list. I am wondering if Skyline supports MSP format files

view request
Shimadzu 8060RX lcd contains no usable data
(3 responses) wesley vermaelen35866 2024-10-22 22:39

Hi,
I've been using Skyline to process data from our Shimadzu 8060NX for a while now and everything works well.
Recently we acquired a new Shimadzu 8060RX but for some reason when I import the results from this machine an error occurs: "The sample contains no usable data".
When I use labsolutions from Shimadzu there is clearly data in the file.

When I use SeeMS from ProteoWizard there is also no data to be seen.

Do you have any idea what is going on here?

I have included two files, one from 8060NX that works and one from the 8060RX that doesn't work. These are similar analysis.

I am very grateful for your help.
Wesley.

 Shimadzu8060RX.lcd  Shimadzu8060NX.lcd 
view request
List of atom's exact masses and isotope ratios
(3 responses) Juan C. Rojas E. 2024-10-30 08:53

Hi all,

Where can I find the list of exact masses and ratios assigned to each of the isotopes used for Skyline?

I am seeing some inconsistencies with the exact mass determined up to the 6th decimal for some large glycan PTMs and I am trying to find the source of the error.

Sincerely,
Juan C.

view request
Ion mobility now showing and not able to export
(6 responses) laura corveleyn 2024-10-18 02:57

Hi Team,

I am looking at diaPASEF data in Skyline and want to include IM values in the export report. However, the IM values all say "NA" and when I want to show IM on the chromatogram (by right clicking) it doesn't show anything. But when I double click the peak I can see the mobilogram so the ion mobility data is definitely read in correctly. How can I fix this?

Best,

Laura

view request
Ion Ratio Standards VS Sample
(3 responses) sabrina s 2024-10-28 10:07
Hi,

I'm using Skyline version 24.1.0.199 for peptides quantification using only light peptides as external standards calibration curve for quantification. I want to add in the report the Ion ratio to compare the ion ratio found in sample and in standards calibration. Is it possible to help me?
Thank you,
view request
Intensity values for quantification
(1 response) cvadival 2024-10-28 12:52

Hello,
We are using Skyline for quantification, but we're unable to find the intensity values; we are only getting the retention time values. Please see the attached pictures for reference. Kindly assist me to get actual intensity value for quantification.
Thanks
VC

 Screenshot_1.png  Screenshot_2.png 
view request
One Peptide Identified in One Replicate but Not in Others
(10 responses) amirhossein abbas-nia 2024-10-25 08:11

Hi
I'm facing an issue with the detection of the peptide "R.ALLNAIGNLELTGAFAEALK.N [146, 165]" across different replicates in my analysis (photos attached). This peptide is successfully identified in replicate 1_RA1_01_2682, but none of the other peptides in this sample are detected, and this same peptide isn't identified in the other replicates.
Interestingly, this peptide wasn't initially identified either, and I'm unsure what I did that led to it being detected. I would really appreciate any guidance on what steps I might have accidentally taken to identify this peptide, as this could help me replicate the process for the other replicates. This is quite important for my analysis.

Thanks so much for your assistance!
Amir

 1.jpg  2.jpg 
view request
Constructing standard curve using isotopoluges
(1 response) sstoychev23513 2024-10-27 14:33

Hi Skyline Team,

Is it possible to construct a standard curve from 5 isotopoluges of a peptide within Skyline? Currently Skyline treats these as 5 different peptides and shows the peak areas individually so it seems only way is to export the total peak area for each of the heavy labelled versions and further process externally from Skyline.

 VVGGL peak area.jpg  VVGGL.jpg 
view request
using DIA-NN for predicted spectral libraries
(2 responses) afshari1 2024-10-21 07:40

Hi,
I am wondering if we can use the “DIA-NN” predicted spectral libraries with Skyline.

view request
Ion Ratio
sabrina s 2024-10-24 11:35
Hi,

I'm using Skyline version 23.1.0.380 for peptides quantification using only light peptides as external standards. I want to add in the report the Ion ratio. Is it possible to help me?
Thank you,
view request
Adding additional peptide after importing results
(2 responses) s jager 2023-09-26 04:46

Hi skyline team!

I just imported my results (.raw files) after loading my fasta, but found out that my fasta was missing a pepitde... is there a way i can add this peptide now and integrate it for the result files? The loading of the files took about 4 hours, so i would prefer not having to start from scratch...

Best Shelley

view request
Drift Time Window Integration
(3 responses) asolosky 2024-10-22 10:29

Hello, my name is Amie and I taught the Ion Mobility filtering for lipids tutorial during the last online Skyline course (last week). During the tutorial, we discussed how it would be helpful to be able to change the integration bounds within the drift time dimension, similar to how we are able to with chromatograms. Is this possible? Normally, we get around this by changing the resolving power window in the ion mobility transition settings, but sometimes there are peaks that fall outside the horizontal violet box that we would like to be able to include in our integrations (i.e., conformers), but only for certain targets, so I wouldn't want to change the filtering for the whole document. That is why I think being able to integrate within the drift time window just like we do for LC would also be super beneficial.

I've included an example data file where this would be helpful. The retention time of interest is 9.6 mins, but at this retention time and m/z there are multiple peaks in the drift window. I would like to be able to integrate the first two peaks at drift times 23.56 and 25.36 ms (summed), but not the peak at 33.96 ms. Can you make this possible in Skyline?

Thanks in advance for taking a look at this!

view request
Following up on importing PD files w/ no q-values
maa98 2024-10-22 18:06

Hi Skyline team,

I tried following the advice on the post below but it didn't fix my problem.

https://skyline.ms/announcements/home/support/thread.view?rowId=42832

My orginal search against my raw data was done w/out a percolator but with fixed value PSM Validator since I was only searching against one protein. I received the same error message as that given in the post above but unlike them, I couldn't get the warning to go away even after setting my score threshhold to 0. Am I filling out the wrong part of the form? Thanks!

Cheers,
Martín

 Screen Shot 2024-10-22 at 6.04.50 PM.png  Screen Shot 2024-10-22 at 6.04.58 PM.png 
view request
different heavy peptide from light peptide for quantification
(1 response) 1638393444 2024-10-21 20:06

Dear skyline supporter, I am doing PRM to quantify my protein A. My heavy protein B is not the same as light protein A. B and A has many similar peptides, so I want to use one peptide of heavy protein B to quantify one peptide of protein A. How can I show them together with chromatograph?

view request
Distinct combinations of heavy isotopes as distinct precursors rather than deduplicated into the same
(1 response) jamie o brett 2024-10-21 16:57

For small molecules containing two heavy tracers with similar heavy-added mass (e.g., C13 and N15, or C13 and H2), it seems that Skyline does not recognize as distinct any adducts that contain combinations of these tracers that have highly similar m/z. For example, for glutamine as the small molecule, the following transition list only gives a single M2C13-H transition instead of a separate one for each row. For high-resolution instruments we know it is possible to discriminate these. Is there a way to import a transition list into Skyline so that these appear as separate precursors?

Transition List (of 3 distinct precursors with very slightly different m/z):
Molecule Name Molecular Formula Precursor Adduct Precursor Charge
Glutamine C5H10N2O3 M2C13-H -1
Glutamine C5H10N2O3 M1C131N15-H -1
Glutamine C5H10N2O3 M1C131H2-H -1

Resulting Targets List (just 1 precursor)
Glutamine: precursor: 147.0686[M2C13-H]

 GlutamineExample.png 
view request
Extract specific mass/charge ratio ions
(5 responses) sunxinyi 2023-05-24 05:31

Hi!
I'm working on a new label similar to TMT(tandem mass tag) applied to proteomics. Now I need to extract the quantitative information of the label (with a specific mass/charge ratio) from the secondary spectrum of the peptide. I wonder if skyline can do that. If it can be done, please tell me how to do it.
Be grateful!

view request
Ion Mobility Resolving Power Units
(4 responses) lauren royer 2024-10-16 11:57

Hello,

When setting the Resolving Power under Transition Settings > Ion Mobility, does this always reference the RP in milliseconds?
I know when searching for a targeted value, Skyline prefers CCS over arrival time when available. I've tried creating a library with only CCS and another with only arrival times to test this but the resulting bounds of the filtered mobility bands appear to be the same.

Thank you!
Lauren

view request
Metabolites precursors: How to filter multiple hits when only 1 retention time detected.
(3 responses) waltteri hosia22589 2024-10-15 06:23

Hello, back to Skyline after some time. Now testing metabolites precursor only HR workflow -with glycans, but question applies to metbolites in general I believe. Is there a way to filter so that skyline would choose only the nearest by RT in transition list as a hit, and discard other isobaric hits in transition list, and results -when there is only one retention time present?
With metbolites, and glycans in particular, isobaric structures are separated by retention time. Skyline picks all isobaric close-in-retention time compounds as hits despite they share the same RT/EIC. Restiricting retention time too tight dosen't work because retention times in transition list are predictions and because they can wobble a bit in hilic. In the attached screenshot Skyline marks all isobaric glycans to be there, G0, A2 and A1B, showing same EIC three times. The glycan nearest to predicted RT, A1B, is the only one actually there.

 nearest RT match.JPG 
view request
Is it possible to build a library on Skyline from a library generated on FragPipe?
litielecruz 2024-10-21 08:32

I generated a DDA library on FragPipe (.tsv), and now I want to use this library on Skyline to analyze some DIA data.
I had no success. Is there any way to do that?

view request
Links to Protter and SRMcollider not working
stolltho 2024-10-17 20:37

Hi there,
I'd like to use tools Protter and SRM collider, but links to both tools are 'dead'.

Clicking on Protter in Skyline leads you to this
https://wlab.ethz.ch/wollscheidlab/#file=5fa311d8-5fea-4770-9f45-d80a65716917.csv

The link to SRM collider returns a blank page, is SRM collider still active?

Cheers,
Thomas

view request
Downloading human proteome?
(3 responses) ayacoob 2024-10-17 08:54

Hi!

I was asked to download the human, mouse, and bovine proteome (reviewed) to load into Skyline. I've used programs with spectral libraries before, like Bruker's XRD spectral libraries, Thermo's FT-IR databases, etc... Usually they're pre-loaded or my lab purchased them for a few years, but I'm still new to Skyline and haven't really done this before.

My instrument is a Thermo Scientific TSQ Altis (QQQ), and I just needed some help on trying to understand where I go to find these databases/libraries so that I could download it and upload to Skyline and presumably use the library to extract/detect as many proteins as I can in a sample.

The best I've done is using the PRIDE Peptidome website and trying to download their spectrum libraries, but I'm not able to open or import the .msp.gz files into any program. I'd be grateful for any help!

Thanks!

view request
Concurrent inclusion of light and heavy fragment ions from imported transition list
(4 responses) diaz-galiano 2024-10-16 07:00
Dear Skyline Support Team,

I am writing to report an issue I encountered when importing a .csv transition list containing light and heavy fragment ions.

Specifically, for certain compounds like labelled internal standards or halogenated compounds, I would like to monitor isotopologue fragment ions. In my .csv file, I have distinct entries for each species. For example, for 13C4-PFOA_H in the attached file (lines 272 and 273), I include the fragments C3F7[M-] (168.98937) and C'3F7[M-] (171.99944). However, when importing the transition list, only the light fragment is added to the target list, while the heavy fragment can only be added manually within Skyline. Could you confirm if this is a bug or if I am missing a step in the import process?

Additionally, I would greatly appreciate guidance on how to properly indicate adducts for heavy species within the same compound name. For example, for PFBA (lines 5 and 7), I can easily define [M-H] and [M-COOH] adducts directly from the molecular formula. However, for heavy species (lines 264 and 265), I have been naming them _H and _COOH, as I haven't found the appropriate nomenclature for indicating heavy isotope losses. Is there a recommended way to label these, e.g., something like [M-(13C)OOH], to display both precursors under the same compound?

I am using Skyline (64-bit) 24.1.0.199 (6a0775ef83) on Windows 10 64-bit. I have attached the .csv file used to create the target list.

Thank you for your time and assistance.

Best regards,
Francisco José
 pfas_complete_precursor_01.csv 
view request
dia-PASEF and spectral library
mvm35 2024-10-17 00:50

Is a spectral library also required for dia-PASEF as is for DIA data acquisition?

view request
Error
(1 response) anhdao darcy 2024-10-16 12:58

Hello,

I observed the red X in the target panel What does it mean and how can I fix it? Please see the attachment t for more information.

Thanks

 Slykine.pptx 
view request
Can I export heavy peptide areas in Report?
(2 responses) martinnm 2024-07-27 18:27

Hi Skyline Team,

I want to export the normalized peak areas of heavy and light versions of my target peptides in my Report. I know I can customize my report and include "RatioLightToHeavy" but I was wondering if there is a selection for the normalized area specifically for the heavy peptide?

Thanks!

view request
SKYLINE sensitivity vs Waters MassLynx
(2 responses) sesharma 2024-10-15 11:23

Compound 1 uses the same parameters and at the same concentration shows:
Analyte area in MassLynx: 38511.961
Analyte area in SKYLINE: 2356949

Compound 2 uses the same parameters and at the same concentration shows:
Analyte area in MassLynx: 30827.227
Analyte area in SKYLINE: 1851899

I am experiencing the above discrepancy. Would you suggest any possible explanation as to why the data analysis through SKYLINE is 100 folds more?

view request
More significant figures possible in peak boundaries report?
(2 responses) sstewart 2024-08-21 17:08

Hi team,
Thanks so much for all these new command line additions!

Is there a way to set number of sig figs in exported reports? My goal is to get higher precision peak boundaries via the command line. I can right click on the column in Document Grid and set to Full Precision, but this isn't sticky. Precision resets to 2 decimal places after I close Document Grid.

Right now, peak boundaries are truncated after two decimal places and this is enough (.06 seconds) to be a noticeable wiggle.

Judging from line 3358 in CommandLine.cs, it looks like number of sig figures might be persisted in viewInfo? I'm happy to submit a patch if you can give me a rough outline of what to do, or maybe I'm just missing something obvious in the GUI.

Thanks as always,
Sam

view request
Koina error running process
(1 response) giocanil 2024-10-15 14:31

Dear Skyline developers,

I am encountering an error while running Koina for EncyclopeDIA library generation.
The error is attached, thank you very much for support.

Giovanni

 koinaError.txt 
view request
Skyline version 24.1.0.199
(1 response) anhdao darcy 2024-10-15 09:17

Hello,

I have just downloaded a new version of Skyline. However, in the "peptide setting" section of the new version, the normalization method shows no option of "Ratio to Heavy." Do you know why? Please see the attachment for more information.

Thank you
Anhdao

view request
Online courses
anhdao darcy 2024-10-15 09:21

Hello,

Please let me know when the next online courses for beginners will be.

Thanks
Anhdao

view request
Finding targets based on isotopic pattern
(1 response) j k o nygren 2024-10-10 07:23

Hello Skyline team,
I'm new to this software and I'm trying to figure out if it can help me with what I currently need to do.
After going through some tutorials, webinars and questions in the forum, I thought it better to ask if this is possible.

I am analyzing polychlorinated alkanes (PCAs) using direct injection and HRMS data.
A common way of doing this with HRMS data is to, for each congener, extract the two (or three) most abundant m/z of the characteristic isotope clusters of highly chlorinated compounds.
For example for C14H20Cl10, that would be [M+4], [M+6] and [M+2], see attached image.
The relative intensity between the most abundant m/z and the second (and third) most abundant m/z are used to identify the congener.
The most abundant m/z is then corrected by the injection standard and used for quantification.

What I hope that Skyline can help me with is to match the theoretic isotopic pattern to the experimental isotopic pattern, and integrate the potential EIC on matches.
So far, I've learned how to get matches based on masses but not on isotopic patterns of chlorine. Is that possible and if yes, are there some functions/tools that I am missing?

For a deeper explanation, see the materials and methods section of Bogdal et al 2015: https://pubs-acs-org.vu-nl.idm.oclc.org/doi/10.1021/ac504444d#_i2

Best regards,
Jonatan Nygren

 example_IsotopicDistributionC14H20Cl10_fromChemCalc.png 
view request
Why is the "Agilent Automation" feature not available on my skyline operation panel?
(1 response) 2378572477 2024-10-09 01:21

Hello

"I want to optimize CE using 'Automation' as shown in the figure."

But why doesn't my operation panel have 'Automation'?

"Please see the attachment."

Thanks,
Zhaoxuanxuan

 问题.docx 
view request
How to Combine Skyline and Mass Hunter:
z15839328872 2024-10-09 01:33

Hello, dear Skyline team!
I want to optimize collision energy using Skyline now. I saw a tutorial that said it could be used with Masshunter, but why can't I use it? Here's a screenshot of my problem.
Zhouyanan

 1cf8a69077bad72e332ed054e3e7703.png  0b495051cb27bcf1a12a2aa9490ddb8.png  681a0f3f969f710450b66fd3bacdf5f.jpg  skyline 使用——筛选合适的肽段和离子.pdf 
view request
How to use Skyline to optimize MRM parameters
(2 responses) z15839328872 2024-10-08 01:48

Dear Skyline Team,
I'm having some issues with Skyline to optimize MRM parameters, and I don't know how to automatically optimize collision energy and declustering voltage. Since my previous protein and peptide identification was done by sending samples, using Thermo Fisher Fisher Scientific's fluid, the raw data was imported into Skyline and the ion pairing information was exported. Later, I used my own Agilent instrument to verify that the peak was too low, and I wanted to optimize the MRM parameters, but I haven't found how to use Skyline optimization, do you have any good suggestions? Or am I using a 2023 version that needs to be updated?
Zhou Yanan

 69efae6c4496f44cc462b4e13cf997f.png 
view request
Export MRM to Waters TQ CE formating should be "eV"
(5 responses) per larsson 2024-09-25 10:51

Version:Skyline-daily (64-bit) 24.1.1.254 (ba32f7f50)
I have had problems to export MRMs to .exp file for waters TQAbsolute. I was able to find the cause. I used a template file with an MRM created in Masslynx.

Skyline writes the "CollisionEnergy(V)_10.000," (for a collision energy of 10eV) this results in an error in Masslynx since collision energy is not regognized.

In masslynx files the collision energy is written as "CollisionEnergy(eV)_1,10.000" the missing e results in that the CEs are not read by masslynx. This can be fixed in word by replacing the wrong term with correct but it would be better if skyline developers fixed this.

Also please make it possible to export retention times (start and stop) and polarity from skyline to masslynx

Regards,
Per

view request
importing multiple complex crosslinked peptides
(3 responses) sri ramarathinam 2024-10-02 21:49
Hi
Thanks again for making such a fantastic tool as Skyline!
I might be missing something, but I was unable to import crosslinked peptides. I was trying to replicate what was done in following ASMS presentation https://skyline.ms/wiki/home/software/Skyline/page.view?name=Crosslinking#:~:text=Support%20for%20crosslinked%20peptides%20was%20first%20added%20to%20Skyline%20version

In this poster there is a dialogue box that says add crosslinked peptides (screenshot attached)- I am not sure how to get to this screen and specifically how to add multiple peptides using the String based representation. If I try via Edit>insert Peptides it recognises "-" as invalid character so I am assuming this is someplace else.

If I have over 100 crosslinked peptide combinations (2 or 3 possible links) and can generate a text output, would be nice if I could import them into Skyline all at once in format that was demonstrated in the poster:
PKEPTIKDEA-PEKPTIDKEB-PEPKTIDEKC[+57.021464]-[+138.06808@2,3,*][+138.06808@7,*,4]

Many thanks again
Sri
 crosslinked_peptides.png 
view request
How to develop MRM method for Metabolites
(1 response) Drymoglossum 2024-10-02 05:35

I have untargeted DDA data for metabolite samples and have used public .msp libraries from RIKEN for MS-DIAL data analysis.
I have access to .raw files of the samples & .msp libraries.
I identified over 100 differentially abundant metabolites that I want to further investigate.
How can I develop an MRM method for these metabolites?
Does anyone have a latest tutorial or guidance on how to approach this? I’ve heard that Skyline might be useful, but I'm unsure where to start. Any advice would be appreciated!

view request
Spectral Library build Issue
(1 response) whdgka75 2024-10-01 18:54

We are trying to build a library using raw files of mass spectrum data contributed in other existing studies.

But as the skyline is updated, some problems arise.

During the initial file upload process, RAW format files cannot be uploaded. (picture 1)

At this stage, only ".dat" file formats are read, and ".mzid" and ".mzxml" files are not readable.

At this stage, may I know if there are any file formats that can be uploaded?

If only a specific file format is possible to uploaded, I would like to know if the library cannot be built if the contributed mass spectrum data does not have that file format.

We built the library by uploading the ".dat" file in the first step and registering the ".raw" file in the next step.

In the next step, enter “search condition”, and under “modification” I want to select carbamidomethyl(M).

However, carbamidomethyl(M) is said to already exist, but cannot be found in the list. (picture 2)

In this case, carbamidomethyl(M) is not in the list because it is a modification that is applied by default, and I would like to know if it is applied automatically or if another error occurred.

 picture1.png  picture2.png 
view request
Unscheduled PRM data analysis
(1 response) afshari1 2024-10-01 11:18

Hello,

I have generated an unscheduled PRM method using Skyline for 200 precursor ions. Could you please provide a guide on how to analyze the resulting PRM data in Skyline? Moreover, Is it necessary to import spectral libraries for this analysis?

view request
Lost my points across the peak visual in chromatogram viewer
(3 responses) Lindsay Pino 2024-09-30 19:02

Hi all :)

I seem to have lost my little red scan number demarkations with the small red number that indicated how many scans were across my peak when I was on the Transition Targets level to view my chromatograms. I attached a minimal doc that I think captures the issue, I'm sure I'm just missing some Magical Checkbox in the settings!

If you need another other information or details, please let me know

Thanks,
Lindsay

 nopoints4skyline.sky.zip  skyline screenshot.png 
view request
PFAS-Specific adducts not uploading in transition list
(4 responses) karamj33405 2024-09-27 12:41
Hello,
My name is Kara Joseph with the Baker Lab at UNC Chapel Hill and we are working on expanding our library for PFAS with some new adducts, however the adducts listed below cause errors when inserting the Transition List. I have included the original adducts we prefer to use below. Do you have any suggestions on how to format these while retaining the structural information for the losses?

M-COOH-HF
M-C2H4-OH
M-CH2-COOH

The only way I can get the Transition List to import properly is as follows:
M-C-2H-2O-F
M-2C-5H-O
M-2C-3H-2O

Thank you!!!
view request
DDA (eclipse orbitrap)
(1 response) zeinab mokhtari31455 2024-09-25 13:09
view request
No pressure traces in Thermo Fisher Files
(2 responses) sascha 2024-09-24 07:45

Dear Skyline people,

I am trying to extract pressure traces from our Thermo Fisher raw files recorded using a Q Exactive couples to a Dionex Ultimate 3000. However, I am not seeing any pressure traces, although they are stored in as "Pump Pressure" within the file (verifyable using Freestyle). It works for files recorded using a Vanquish HPLC.

Can you help me find out what's going wrong?

I attached you example raw files as well as an example document.

Thank you very much for the help!

 QC.sky.zip  QC_Dionex.raw  QC_Vanquish.raw 
view request
Skyline-MRM transitions on QqQ with 2 scan segment issues
(2 responses) weigandm 2024-09-23 06:33

Hello,

I collected MRM transitions on an Agilent QqQ. I had 2 scan segments. One segment was from 0-1.1mins, in which solvent was diverted to waste. The second segment was from 1.1mins to the end in which solvent was diverted to MS for data collection. When I import the results into skyline it only shows that first segment from 0-1.1 mins. I want to see the data from 1.1mins to end which is when analytes in the sample will have abundant MRM transitions. How do I get Skyline to be compatible with this data?

view request
In silico digestion
(1 response) mvm35 2024-09-24 08:19

For in silico digestion using various enzyme.

What do the following mean (X= some amino acid):

[X | P]
[X | - ]
[- | X]

thank you.

view request
Spectral library generation
(2 responses) mlazear 2024-09-23 12:39

Hi,

Is there a way to construct a spectral library such that Skyline will just "trust" the user that parameters are correct? I am trying to build some empirical (e.g. constructed from DDA) spectral libraries for DIA runs where the peptides have exotic/unlocalized modifications - and I am really struggling with how to get Skyline to accept a spectral library.

Ideally, I can just pass in a list of arbitrary ID, precursor m/z, precursor charge, product m/z, library intensity and have Skyline accept it without any "error" checking (e.g. do not try to match the product m/z to the peptide sequence). Is there a way I can fool Skyline into accepting such a list? I don't want to have to manually add the modifications to Skyline, since I have 1000s of them. I am happy to manually construct a library in whatever input format is needed.

view request
Top Down Proteomics data
PP 2024-09-23 13:21

Hello,

I wanted to know if Skyline is able to process top down MS data? If so, is there a tutorial for it?

view request
2024 Skyline Online Course recording
(1 response) skyfall 2024-09-20 08:31

Dear Skyline team,
I won't be able to attend the upcoming online course and was wondering if you will record it? In case yes, would I still need to sign up for the course to watch it?

Thanks a lot
best wishes

view request
Safety of Proprietary Sequences
(2 responses) seamus kelley 2024-09-19 08:17

Hello, I'm interested in adding Skyline to my group's proteomic workflows. We are working with the proprietary sequences. Does your software protect the proprietary nature of these sequences?

Thank you!
Seamus Kelley

view request
Help with Small Molecule Library Building, etc.
(13 responses) mwmann 2024-09-10 10:19

Hi, I'm experimenting with small molecule work in Skyline and have a few questions that I have not seen addressed in the documentation or other questions.

  1. When building spectral libraries, right now I am converting a zenoTOF wiff2 file into mzml, and selecting scans manually for entry into an ssl file. This works, but has been very tedious. In the documentation, there is a list of supported search engines, but they all seem to be for proteomics. Is there a supported search engine for small molecules to aid in library building/annotation? If not, are there any faster ways that would be useful for me to be aware of?

  2. With the above method of creating library entries, I often have high-abundance precursor peaks in the library spectrum. Ideally, I'd like to filter these out from quantitation/selected transitions automatically. The 'Precursor m/z exclusion window' under Transition Settings > Filter seems like it should do this, but in practice it has not worked to filter out these peaks. Can you advise?

Thank you for any help you can provide!

Sincerely,
Morgan Mann

view request
Merge two SRM together
(5 responses) nbekhti 2024-09-10 14:07

I have under my skyline file one same metabolite showing in two lines, and I would like to add the second transition into one same SRM
I tried adding transition, copy paste from another skyline file where they show up together, but the second transition is gray with no data when I do this.

How to fix that issue please?
Thanks
Nihel

 SRM dublicates.JPG 
view request
Replicate comparison - collision energy optimization
(1 response) a b heijnis-2 2024-09-17 04:07

Hey! For our proteomics project we are using the automation tool to connect Skyline with Masshunter from Agilent. Therefore we have two questions.

Recently, we have compared the export methods that we made by going through the steps manually and doing it with the automation tool. Luckily, the results were completely the same!
However, Skyline comes up with the starting point for CE optimization, using step size and step count. We would like to further optimize with the results of the first experiment as an starting point. How could we approach this?

Secondly, when processing our results in Skyline in the replicate comparison, the different CE are always displayed as step (-2, -1, 1, 2 etc.). See figure in the attachments.
We would like to know whether it is possible to have the CE settings shown in de legend as there are many differences between peptides.

Thank you in advance!

Alexander

 Question Skyline.png 
view request
Trouble with Target Quantification and Integration in Skyline for PRM Data
(5 responses) afshari1 2024-09-13 14:44

Hello,

I am analyzing my PRM results using Skyline and encountering an issue with quantifying certain targets. Specifically, in the attached file, there is a peak at 69.x minutes in the bottom panel, which corresponds to one of my targets. The MS/MS data from Thermo Fisher software confirms this match. However, when I import the same RAW file into Skyline, this target is not being quantified or integrated. Could you please help me understand why this might be happening and how I can resolve this issue?

 IMG_0606.jpeg 
view request
Wrong XL transition mass?
(4 responses) tjiang 2024-09-12 08:19
Hi, I'm working on generating library and transition list for XL experiments using DSBU as crosslinker. I had some issues with XL peptide mass.

For example, for the XL peptide "AKGILTC[+57.02146]R - EIPLKVLVK -[+196.0848@2,5]", the correct m/z value for the +4 charge state is 538.824, as shown in the attached "peptide_in_datafile.PNG". However, it shows as 485.3038 in the spectra library as in the attached screenshot named "lib.PNG". And in the transition list, it only gives the XL peptide without cysteine alkylation, as shown in the attached "in_transition_list.PNG".

Could you please help? Thank you so much!
 peptide_in_datafile.PNG  lib.png  in_transition_list.PNG 
view request
Help with SureQuant data analysis
(1 response) darora 2024-09-12 09:03

I´m missing the heavy fragment ions for some of my peptides from a SureQuant run. I can see that these were triggered ( precursor traces)but no fragment ions, unfortunately. These fragment ions were there in the method used for the run. So I´m a bit confused.

view request
How to update KEGG/CAS/HMDB fields?
(1 response) Will Thompson 2024-09-11 06:48

Hi team,

I have a working document for small molecules (attached) and I'd like to use the document grid to add information into the KEGG/CAS/HMDB fields. However, these fields are grey in the document grid and not editable. Please advise on how to batch-enter values into these fields in a manner such as copy/paste from excel, or importing an annotations file?

Cheers

Will

 P148_ACE_v0p5.sky.zip 
view request
Error importing MaxQuant msms.txt file
(2 responses) SChen 2024-09-11 08:42

Dear Skyline team,

I encountered an error while trying to import peptide search result (MaxQuant msms.txt file) into Skyline (see attached error message). I have tested this in Skyline Daily 23.1.1.520 & 24.1.1.254 as well as Skyline 24.1.0.199 and got the same errors from all.

I uploaded both mqpar.xml and msms.txt file to https://skyline.ms/files.url. Because the msms.txt file is quite large (50 GB), I uploaded a smaller version (only the top 10000 line in the original msms.txt). Since I encounter the same issue with the smaller version, hopefully you will also be able to reproduce the error on your end. I also had to change the mqpar.xml file name because there is already a file named after that in https://skyline.ms/files.url.

Please let me know if I can provide more information about the issue.

Thank you in advance!
Shimin

 Skyline error.txt 
view request
Failed to attempt reintegrate peaks_index 6 must be between 0 and 5
(2 responses) vanyabangera 2024-09-10 22:45

Dear Team,

We are failing to train mProphet model in skyline version 22.2.0.527 (841287d47) and 24.1.0.199 (6a0775ef83). But unable to reintegrate. It's showing pop-up "failed attempting to reintegrate peaks. The index 6 must be between 0 and 5". As per the suggestion obtained from previous queries we tried to perform the rescoring but failed to reintegrate. Kindly advice to resolve the issue.

Thanks in advance.

Kind regards
Vanya K N

view request
Skyline command line error "report does not exist"
(6 responses) sstoychev23513 2024-09-10 06:00

Hi Skyline team.

We are setting command line based data extraction where a custom report is meant to be exported at the end of the analysis. When trying to export a standard report template everything runs fine but we get an error when we try to export a custom report. I have attached some sreenshots. Please advise further,

Thank you

Stoyan

 image (6).png  image (7).png  image (8).png 
view request
Failed to attempt reintegrate peaks_index 6 must be between 0 and 5
vanyabangera 2024-09-10 22:48

Dear Team,

We are failing to train mProphet model in skyline version 22.2.0.527 (841287d47) and 24.1.0.199 (6a0775ef83). But unable to reintegrate. It's showing pop-up "failed attempting to reintegrate peaks. The index 6 must be between 0 and 5". As per the suggestion obtained from previous queries we tried to perform the rescoring but failed to reintegrate. Kindly advice to resolve the issue.

Thanks in advance.

Kind regards
Vanya K N

view request