I would need to see your .raw files to tell you why your MS2 chromatograms are not showing up.
Files which are less than 50MB can be attached to these support requests. You can always upload larger files here:
https://skyline.ms/files.url
When the MS/MS Acquisition Method on the "Full Scan" tab at "Settings > Transition Settings" is "DDA", an MS2 spectrum will be included in the extracted chromatogram if the isolated m/z is within the extraction window for the MS1 chromatogram. That extraction window is controlled by the "Mass Accuracy" setting in the "MS1 filtering" section on that Full Scan tab. For this reason, yes, it is possible that the 10ppm mass accuracy that you have specified is preventing any DDA MS2 spectra from being included in the extracted MS2 chromatogram. If you change that to a higher number and then use the "Reimport" button at "Edit > Manage Results" you might see some MS2 chromatograms.
Skyline is not a very good tool for looking at MS2 spectra in DDA experiments. Skyline is very good at extracting chromatograms, but, with DDA data, the same precursor is not sampled at reliable intervals so DDA MS2 chromatograms tend to have long straight lines connecting the times where a matching MS2 spectrum was found.
-- Nick