Hello,

I am hoping to figure out the best column to add to my document grid to display the value of the surrogate standard area used in normalization for each replicate. If there is a simple way to display this unique value for each replicate that would be great.

Thank you,
Zach

 
 
Nick Shulman responded:  2025-09-06 14:04
If your surrogate standard has only one precursor charge state, then the "Total Area" column will contain the number that is being used as the divisor when normalizing against that surrogate standard.

That is, if you select your surrogate standard's Precursor in the Targets tree and then use the "View > Live Reports > Results Grid" menu item, the "Total Area" column will contain the surrogate standard's area for each Replicate.

Does that answer your question?
I might have misunderstood your question, and it might be easier to answer it if I could see your Skyline document.

In Skyline you can use the menu item:
File > Share
to create a .zip file containing your Skyline document and supporting files including extracted chromatograms.
Files which are less than 50MB can be attached to these support requests. You can always upload larger files here:
https://skyline.ms/files.url
-- Nick