It might be helpful if you could send us your Skyline document.
In Skyline you can use the menu item:
File > Share
to create a .zip file containing your Skyline document and supporting files including extracted chromatograms.
Files which are less than 50MB can be attached to these support requests.
You can always upload larger files here:
https://skyline.ms/files.url
I see that you have set "Standard Type" to "Surrogate Standard" in that first screenshot.
You would typically only do that if there is another molecule in the document whose Normalization Method is "Ratio to <that surrogate standard>".
You can read about Surrogate Standards here:
https://skyline.ms/wiki/home/software/Skyline/page.view?name=Surrogate%20Standards
There are two different ways of doing absolute quantification in Skyline.
One way is to have external standards of known concentration. You use the "Analyte Concentration" and "Concentration Multiplier" fields to tell Skyline what the concentration of your analyte is in the external standards (i.e. calibration samples). Skyline tells you what the concentration is in your unknown samples by comparing the signal in the external standards with the signal in the unknown samples. If you are using an internal standard for normalization, the internal standard needs to be spiked in at the same concentration in both the external standards and the unknown samples.
In this case, the actual concentration of the internal standard is irrelevant because it appears in the denominator of the normalized area in both the external standard and the unknown samples.
The other way to do absolute quantification is without any external standards and just an internal standard. In this case, you would tell Skyline the concentration of the internal standard by specifying the "Internal Standard Concentration" value. In this case, the Calculated Concentration ends up being the Normalized Area times the Internal Standard Concentration value (where the "Normalized Area" is the ratio of the Light area to the Heavy area because of the "Normalization Method" chosen on the Quantification tab at "Settings > Peptide Settings").
If there are any external standards in your document then Skyline assumes you want to calculate concentrations using the calibration curve so the "Internal Standard Concentration" value ends up being ignored.
I am not sure I understand your original question. It might make more sense after I see your Skyline document.
-- Nick