Why do you think you need to convert your .raw files to .mzML?
Skyline understands all of the same raw file formats as MSConvert with the exception of Bruker .yep files which Skyline does not include support for in order to save on download size.
If you are using .mzML files, and if your settings on the Full Scan tab at "Settings > Transition Settings" say "Centroid", then, yes, the .mzML file does need to be centroided.
The reason for this is that Skyline wants to use the vendor (i.e. Waters) centroiding algorithms, and those algorithms are only available when reading the original raw data.
So, if you want to use .mzML files with Skyline and you want Skyline to be using centroided data then MSConvert needs to have been told to centroid the data.
The way to tell MSConvertGUI to centroid the data is to choose "Peak Picking" in the "Filters" box and then press "Add".
If you want to use .mzML files that have not been centroided then you could choose a different sort of Mass Analyzer on the Full Scan tab such as "TOF".
Alternatively you could keep that as "Centroided" and tell Skyline to extract chromatograms from the original Waters .raw folders instead of mzML.
If you are still having trouble you can send us your files.
In Skyline you can use the menu item:
File > Share
to create a .zip file containing your Skyline document and supporting files including extracted chromatograms.
Files which are less than 50MB can be attached to these support requests. You should probably send us your .mzML files and your .raw folders and that .sky.zip file. Those will probably be much bigger than 50MB so you can package everything up into a .zip file and upload it here:
https://skyline.ms/files.url
-- Nick