Trimming trnsitions to top ranked

support
Trimming trnsitions to top ranked Zac  2021-10-20 07:17
 

Is there a quick way to refine the number of transitions used for PRM quant. I want to only use the Top 4 ranked transitions per precursor? Thanks

 
 
Nick Shulman responded:  2021-10-20 10:15
If your Skyline document has a spectral library in it, you can go to:
Settings > Transition Settings > Library
and change the number for "Pick X product ions".

If your Skyline document does not have a spectral library, you can create one using the menu item:
File > Export > Spectral Library
this will create a .blib file where the "spectra" consist of the m/z's of the transitions in your Skyline document, and the intensities are the integrated peak areas of your chromatograms.
After you have created a spectral library from your chromatogram data, you can go to:
Settings > Peptide Settings > Library
and push the "Edit List" button, followed by "Add", and then point Skyline at your new .blib file.
And then make sure that your new library is checked in the Peptide Settings Libraries checked list box.
After you have done all that, you can go to:
Settings > Transition Settings > Library and change the number for "Pick X product ions".
Skyline will then choose the X product ions that have the highest intensities in the spectral library.
(I would have thought that there was an easier way to chose the top X transitions without having to export a spectral library, but I cannot find it if it exists)

Note that for each precursor in your document, Skyline remembers whether you have manually changed the set of transitions that the precursor has. You can see in the Document Grid which precursors still have their transitions settings managed by Skyline. The column which controls that is "Auto Select Transitions".
If you want to tell Skyline to automatically manage the transitions for all of the precursors in the document, you can go to:
Refine > Advanced
and check the checkbox for "Auto-select all Transitions".

-- Nick
 
Brendan MacLean responded:  2021-10-20 17:10
If you want to do this based on measured data, you can also use Refine > Advanced - Results - Max transition peak rank as discussed in the Targeted Method Refinement tutorial (bottom of page 18):

https://skyline.ms/tutorial_method_refine.url

This would keep you from having to use File > Export > Spectral Library in the way Nick mentions. It is a bit more direct.

Even though the Targeted Method Refinement tutorial uses SRM data, it may be a good one to do. Much of it would still apply to PRM.

Thanks for posting to the Skyline support board.

--Brendan
 
Zac responded:  2021-10-21 09:14
Thanks... I had gone into the Refine>Advanced-Results but had missed the Max transition peak rank, this is exactly what I need. Appreciate the feedback