Hi Laura,
Yes, I think you will want to use Ion Mobility Filtering. I am not aware of an instrument which pre-filters its spectra based on drift time. The drift time instruments I know of are those from Agilent and Waters. The Bruker timsTOF allows pre-filtering by 1/K0 in a prmPASEF method, and the SCIEX SelexION and Thermo FAIMS allow filtering by Compensation Voltage. If you think, however, that your transition list can be provided to your instrument in its method in a way that will cause it to produce filtered spectra, then in that case using IMS Filtering would be redundant and of no real benefit.
When the CCS values are present in the library, then Skyline will use those preferentially over the drift time values. You should be able to manually edit the values in the grid, but keep in mind: 1) if you edit a drift time value, you should clear the CCS that was calculated from it, or your editing will not impact the filtering, 2) if you edit a CCS value, you should probably clear the original drift time or provide a new drift time to keep from storing out-of-sync values, though in general this is less important than #1 because Skyline will use the CCS value over the drift time value.
Hope this is helpful to you. Thanks for using the IMS features and posting to the Skyline support board.
--Brendan