Hi Jiny,
I'm not a Skyline developer but think I have an idea what the problem might be. The dotted chromatogram lines indicate that the transition is not set as quantitative. I believe that is the reason why the calibration curve can't be generated. In the Transition Settings window and Full-Scan tab, if you have the acquisition method set to DDA, the product ions won't be set to quantitative.
To fix this, you can change the acquisition method to PRM and reimport the data, or set each transition ion as quantitative. To do this:
- Right click on the product ion of interest
- Click on the Quantitative button from the drop down list.
With either method, the extracted ion chromatogram should now be a solid line instead of a dotted line, and your calibration curve should work
Cheers,
Chris