Library missing information

support
Library missing information litielecruz  2023-05-10
 

Hi Skyline team,
I got stuck on a step and I hope someone can help me.
When I build the library in Skyline-daily I realize that I'm missing some information when it is loaded. For example, in a random protein, the DDA (msms.txt) shows it has about 200 peptides found, when I build the library in Skyline using that DDA, Skyline shows me around 100 peptides. The same goes for some modifications that I'm interested in seeing, they are not appearing in the same number that I see in the msms.txt. Does anyone know why I'm losing information when setting up the library? Is it a setting issue? I have tried several setting changes.

Thank you!

Liti

 
 
Brendan MacLean responded:  2023-05-12

Hi Liti,
Can you post your files to:

https://skyline.ms/files.url

Also, it is always helpful to include exact details and/or screenshots. In this case the name of the protein you are using for the example of 200 v 100, and ideally a screenshot showing it in Skyline, or multiple screenshots in a PowerPoint slide deck.

The files we would need are the MaxQuant results you used to build your library and your Skyline file as a .sky.zip (using File > Share).

I think we should be able to give you a quick response once we have these. Sorry for the delay on a response.

--Brendan

 
Brendan MacLean responded:  2023-05-12

Let me also add a bit more on the potential filters that get applied between your starting msms.txt file and what shows up in the Skyline Targets view:

  • In the library build form, you are asked to supply a probability cut-off. Depending on the cut-off you choose, this may filter peptides listed in your msms.txt file.
  • Skyline has a number of peptide-filtering settings that will further reduce the number of peptides appearing in the Targets view from what is stored in the spectral library (.blib file). e.g.
    • Missed cleavages - if you set this to a lower number than you used in your search (e.g. 0) than peptides with a higher number will be excluded
    • Filters: Min length, Max length, Exclude N-terminal AAs, etc. can all filter peptides between the library and the Targets view
    • Modifications: if you are missing modifications that you allowed MaxQuant to use, e.g. Oxidation (M), this would reduce the peptide count
    • Transition Settings: Filters - the precursor charges could cause peptides to be removed, the ion types and charges could combine with the Library settings to remove peptides
    • Transition Settings: Library - if you impose a "minimum product ions" limit, then peptides lacking enough matching transitions may be excluded
    • Transition Settings: Instrument - if the Max m/z value is set lower than the actual range of the data files, this can cause peptides to be excluded. This is important when taking a library from a DDA instrument to a triple-quad SRM instrument, because you don't want Skyline suggesting precursors that the triple-quad cannot measure.

I am sure there are others, but I wanted to point out these filters exist and often need to be tweaked to get everything you want. This has come up before, and I continue to consider if there are better ways for us to handle DDA and DIA search tools when a user wants to use Skyline target everything that was found by an upstream tool, as you seem to want to do.

 
litielecruz responded:  2023-05-16

Dear Brendan,
Thank you for the tips, in fact, I was losing information because of these settings. Mainly because of the missed cleavage numbers. It was set to zero.
I changed it according to the msms.txt result and I have more peptides now.
I will work on this and see if now I have what I need for my analysis.
I come back here if I still have questions.
Thank you again.

Liti

 
litielecruz responded:  2023-05-22

Dear Brendan,
I'm working on my DIA results and a question came up that I'm not sure about.
Considering that all my DDA (library) and DIA acquisitions were in the same instrument and with the same setting e.g. time and gradient for peptides elution, Does the retention times in my DIA results must match with the library data?
I'm asking because I noted that it is not all the peptides that have the best peaks matching with the retention times from the DDA data that the library was created.
I'm wondering if this is important or not and if I need to consider only the DIA peaks in the same retention time from my library.

Thank you again for your help,

All the best,

Liti