oriane strassel responded: |
2023-05-10 |
Update information: I run an SIM acquisition only in positive mode, and I still got the same error when trying to process |
|
|
Brian Pratt responded: |
2023-05-10 |
This is highly unusual. Can you provide a copy of the file? Thanks, Brian Pratt |
|
oriane strassel responded: |
2023-05-10 |
Thank you for your fast answer, I just uploaded the 2 files on the file sharing folder |
|
Brian Pratt responded: |
2023-05-10 |
This is an interesting one! The error happens as we try to read the first chromatogram after the TIC. It's declared as having polarity "mixed" - we're only used to seeing positive or negative. I'm not sure what that would even mean - and there's no way to describe that in mzML - what does it look like when viewed in Agilent software? |
|
Brian Pratt responded: |
2023-05-10 |
In the meantime, msconvert --continueOnError -z D:\data\Oriane\20230510_Biphenyl_OS_MIX1SIM.d should produce a usable mzML file if you don't care about the chromatograms. |
|
oriane strassel responded: |
2023-05-10 |
Hi Brian, Data on MassHunter seems fine (see screenshot attached) Unfortunately I wasn't able to change the file using msconvert, I still got this accessibility issue (access denied) |
|
|
Brian Pratt responded: |
2023-05-11 |
I'm not sure what you mean by accessibility issues, but that sounds like you probably just need to tell msconvert to write to a different folder. Best, Brian |
|