I'm running a small molecule targeted (steroid hormone panel) analysis (LC-MS on a QExactive) using MS1 filtering (surprisingly better results than PRM... a discussion that I'm hearing quite a few users agree). I have spiked deuterated internal standards to my samples. Unsurprisingly, those with higher number of D atoms produce a noticeable shift in the retention times. However, peak boundaries can not be separately set for target and internal standard (setting the peak on one automatically sets the other). Is there any way to overide this feature? I'd like to manually determine the peak boundaries, as Skyline is getting a little confused establishing the peak start-end times. This creates a problem when integrating low abundance samples, as the 'wide' peaks of the internal standards overestimate the areas of the unlabeled precursors.

Thanks!!!

Alejandro
 
 
Nick Shulman responded:  2021-03-23
You can tell Skyline that an isotope modification will have a different retention time than the unmodified peptide.

To change that, edit the isotope modification by going to:
Settings > Peptide Settings > Modifications > Edit List > Edit

At the bottom of the "Edit Isotope Modification" dialog, there is a setting "Relative Retention Time". This defaults to "Matching". I think if you change it to "Preceding", I think Skyline might make it easier for you to adjust your peak boundaries the way you want them. (I actually am not sure exactly what happens when you choose "Preceding". I know that if you choose "Unknown", Skyline will let you adjust the peak boundaries of your precursors completely independently from each other, but I'm not sure what happens with "Preceding").

By the way, you can always adjust the peak boundaries of transitions independently of each other, but Skyline intentionally makes it difficult. If you right-click on the chromatogram and choose "Transitions > Single", then, if you have a single transition selected in the Targets tree, any adjustments that you make to peak boundaries will only be applied to that one transition.
-- Nick
 
Brendan MacLean responded:  2021-03-23
I was going to explain what Nick explained for peptides, but thanks to Nick, I don't have to. There is no equivalent for small molecules that I know of. However, you can always define your standard as a separate molecule and then Skyline will integrate them separately, and by using multiple selection in the Targets view (Shift-click or Ctrl-click) you can get Skyline to plot the two values on the same graphs. It is not quite as convenient as the way we group precursors for the same molecule, but it could suffice as a workaround. It would certainly allow you to get the peak integration boundaries and peak areas you are trying to achieve. If your goal is to get those into a report for statistical processing outside Skyline, that might be enough.

I would expect MS1 chromatograms to work better than PRM in a simple solution. With MS1, you are generally working against other ions, either because they are much more abundant than what you are trying to measure and fill the trap, reducing the precision with which your target molecule can be measured, or by direct interference with the molecule you are trying to measure. You could use SIM to solve the first issue, but only PRM can help you with interference and noise reduction.

We do hear from our pharma partners that they prefer MS1 for low complexity samples. Conversely, ion filtering with a quadrupole or FAIMS device appears to be much more important in high complexity and high dynamic range samples like cell lysates or plasma.

Thanks for posting to the Skyline support board.

--Brendan
 
alejandro.cohen responded:  2021-03-23
Thanks Brendan, Nick

I was about to mention the absence of that setting for small molecules (maybe a suggestion for future releases?).

With respect to Brendan's suggestion (ctrl-click), would the normalization still work? That is, will Skyline know one is the Internal Standard of the other automatically, for calibration curves and interpolating Values of unknowns?

Thanks again, Skyline rocks... I'm creating a small Skyline-sect here at Dalhousie University.

Best

Alejandro
 
Brendan MacLean responded:  2021-03-23
You would not get the automatic light:heavy normalization, but if you had a small enough number of these, you could use the "surrogate standard" mechanism to pair them up manually. Not great. Yes, it seems like an option like the retention time matching for peptides would be useful for small molecules. We just hadn't seen a case of that yet.

Thanks for the feedback.

--Brendan
 
alejandro.cohen responded:  2021-03-24
Thanks Brendan!

I have 18 targets. I created 36 molecules, just as you suggested. I do have some questions:

I did go through the tutorials, I created the Normalization Method Column, and selected ratio to Heavy, but can't find a way to individually link each molecule to it's respective surrogate (I would have imagined another column with a drop down menu linking each molecule to it's surrogate).
The tutorials only give an example of one surrogate.

I cant see the normalized areas, so clearly I'm screwing up.

Attached the file for you to inspect, if you have the time to inspect it.
 
Nick Shulman responded:  2021-03-24
Alejandro,

In order to be able to choose "Ratio to Surrogate Xxx" as a normalization method, you first have to change the "Standard Type" of some other molecule to "Surrogate Standard".
There is an explanation of how the Surrogate Standards feature works here:
https://skyline.ms/wiki/home/software/Skyline/page.view?name=Surrogate%20Standards
-- Nick
 
alejandro.cohen responded:  2021-03-24
WORKED!!! Solved. Thanks N&B!