Mass error and isobaric peaks analysis

support
Mass error and isobaric peaks analysis Vahid F  2019-09-06 08:31
 

Hi all,

I have acquired some MS/MS spectra at 120k resolution in PRM mode on QE-HF to be able to resolve small isobaric ions labeled by one 13C or 2H or 15N. Data files are imported into Skyline and I can see the MS/MS in Full Scan panel (picture attached).

The issue is that due to a ~5ppm mass error, m/z that Skyline is looking at is slightly off. Is there a way to force skyline to look at the peaks and not the theoretical m/z or somehow or set a m/z offset? The attached picture illustrates the issue well.

Thanks,
-Vahid

 
 
Brendan MacLean responded:  2019-09-07 11:22

Hi Vahid,
The setting you are seeking does not currently exist in Skyline. I implemented it once for some poorly calibrated data from a collaborator, but in that case the answer ended up being more complicated than just a single offset, because the mass error varied systematically with the m/z causing curvature when you plotted mass error by m/z:

https://skyline.ms/_webdav/home/software/Skyline/events/2017%20Webinars/Webinar%2015/%40files/Webinar%2015%20%28MacLean%29.pdf#page=25

In the end, the collaborator recalibrated the data outside Skyline and then it worked just file without needing any kind of offset.

I would suggest you pursue a similar solution. There are a lot of tools out there that do interesting things with m/z calibration. Your simple offset request seems like something ProteoWizard msconvert should do, but I am not seeing it in the docmentation:

http://proteowizard.sourceforge.net/tools/msconvert.html

Maybe check with ProteoWizard support:

http://proteowizard.sourceforge.net/support.html

Thanks for using Skyline. Sorry, we don't have a great solution for dealing with miscalibration.

--Brendan