Dear Andrea,
you can process both multiplexed and non-multiplexed PRM data.
In case of non-multiplexing skyline seems to be able to even differentiate ions from light and heavy if they have the same mass. It means that potentially all ion types can be used as long as each precursor is in just 1 isolation window. (This is from a personal quick test, i did not explore this throughout)
In case of multiplexing you need to know which fragment ion types still contain the label mass shift and you should only use those for quantification. In principle you can still use all ion types for visualization and potentially for identification, but light heavy fragment ions of the same mass will be added up and will be of the same intensity for light and heavy.
Please be aware that PRM can result in very slightly different light-heavy ratios than MS1 since it tends to very slightly overestimate the amount of heavy.
For transition settings you might want to allow all ion types and charges you want to build chromatograms from:
Precursor Charge: 2, maybe also 3
Ion Charge: 1,2
Ion Types: p, y (others might be used too but mostly just y-ions are quantitative)
From: Ion 3 (y1 and y2 are not very specific and can have isobaric interferences), fragment ions with a m/z >precursor are more specific so this might be a more stringent option
To: 6 ions or last ion
In Full Scan Settings, MS/MS Filtering set to targeted and input the settings of the acquisition, profile data recommended.
Other settings are (more or less) optional.
If you have a testrun of the heavy labled peptides you can use respective database searches to build a spectral library (or from the experiment results). This allows you to build chromatograms automatically just for the X most intense (most relevant) fragment ions for each peptide.
Hope it helps,
tobi