Hi Jingjing,
For this, we recommend you use "Centroided" as your mass analyzer and then simply use 10 ppm as your mass tolerance. This has worked quite well for us and has even become our preferred mode of extracting chromatograms from Orbitrap data. With this set-up, you will not see any chromatogram peaks with mass errors greater than 10 ppm.
There is currently no way of telling Skyline that you will allow it to extract from a wider range but not allow peaks in that range to be considered.
Hope this is helpful in your data processing with Skyline.
--Brendan