Hi Brian,
Thanks for your answer and document. I have looked at it and do think you have noticed with the "fake molecule" approach, the same issue, meaning inversion of Low CE and High CE in Skyline document.
I have attached some slides to explain what I see.
- Slide 1 and 2 are showing that for "compoundA" (175 ->116 m/z) and "fake_molecule" (116->175 m/z), precursors ions are extracted from high CE scans (even scan number), when products ions are extracted from low CE scans (odd scan number).
This is further confirmed by precursors ions intensity you expect to see higher in MS1 vs MS/MS and products ions intensity you expect to see higher in MS/MS vs MS1.
-Slide 3 should convince us that Low/High CE scans are inverted in Skyline vs Bruker DataAnalysis. Notice the scan numbers, the corresponding MS level and our target precursor/product ions intensity in the mass spectra.
-Slide 4 is looking at the Bruker Raw file with SeeMS and confirming that odd scan numbers are Low CE and even scan numbers are High CE, because of the fragmentation spectra profile that correspond to what we have seen in DataAnalysis.
Please, have a look and tell me if my interpretation seems correct to you?
If yes, I would invert MS level information given to odd/even scan during data import. On the other hand, I do not know if this is a constant data structure in Bruker bbCID file? My knowledge stops here...and I rely on your help.
Thanks again for looking at it.
Best regards,
Christophe