Hi Monica,
Maybe try widening your mass accuracy to 20 ppm. You are extracting from centroided data with 10 ppm mass accuracy. Sometimes when you see abrupt signal drop-off like that it is because the mass accuracy is not stable enough and your MS1 peak of interest is coming and going within the extraction window.
The most extreme case I saw was a large peak in TASEFDSAIAQDK, but the jaggedness was less extreme in the picked peak. That might be just that the jagged peak is not the right peak and truly falls close to the edge of the extraction range, but m/z variance causes its signal to appear and disappear.
I often try a bit higher mass accuracy constraint (e.g. 20 ppm) when I see stuff like this. It would at least be interesting to see the impact of that.
Thanks for the library update. Not sure how you got what was in the ZIP file, it had no spectra at all, but was otherwise a well formed SQLite file, not corrupted on that level. If you can find a way of reproducing that, it is something we would be interested in fixing.
Thanks for your post to the Skyline support board.
--Brendan |