Hi Jens,
Can you post a screenshot of your Transition Settings - Full-Scan tab? If Skyline is ignoring the explicit settings you have made, then the first most likely explanation is that you have (accidentally) told it to do that by specifying your extraction retention times as "Include all matching scans", instead of using "Use only scans within [] minutes of predicted RT", which is the only setting where Skyline consults the explicit retention times you have set.
I can see the problem in this case, since you have also set an explicit window and the specified minutes would seem to override that.
Perhaps we should make "Include all matching scans" respect explicit times, as we still allow Transition Settings - Instrument - Min and Max time to restrict the definition of "matching scans".
You are the first to run into this. So, it may be that we have not thought it through thoroughly enough, but I am still interested in knowing whether you could at least get part of what you are looking for by using the "predicted RT" option in the Transition Settings - Full-Scan tab, or if you are already using that, then we'll need to keep searching for why Explicit Retention Time is getting ignored in your document.
--Brendan |
Hi Brendan,
thanks for your response.
Yes, you are totally correct in assuming that I used the "Include all matching scans" setting. But this was due to the fact that Skyline is asking for a Prediction algorithm when I am using "Use only scans within [] minutes of predicted RT" (see attachment). I then tried to include the list of modified peptides with their RTs in the predictor list, but it is giving me the response, that there are peptides with the same mods (as e.g. deamidation and iso-aspartate result in the same mass shift at the same position) appearing more than once. Is there a way to circumvent setting a RT Prediction algorithm and just use the settings/times from the document grid list?
Thanks a lot in advance!
Regards,
Jens |