Hi Dan,
I will add my $0.02. You are correct that decreasing the resolution of the quad isolation (i.e. increasing the isolation width) will increase the intensity of your targets. This is definitely an old and common trick to improve sensitivity in the absence of interference. It is also partially why DIA has been enabled. While DIA is much less selective, the transmission of a 5 m/z isolation window is much better than a 1.4 m/z isolation window. Think of it as a pipe, the larger the pipe the faster you can move water through it. The same thing is true with quad isolation. You see the increase in signal because of the reduced ion trap fill time as the signal is a function of the current/fill time. So you don't have any more actual signal you just got there in less time.
You're also correct that if you isolate the entire isotope distribution that you will get isotope distributions on the product ions that could give you more ion counts if used correctly. We have been thinking about ways to use the entire isotope distribution like you said to deisotope the spectra and then extract just the signal that is reflective of the entire isotope distribution. Nick Shulman on the Skyline team did this previously in a different tool called Topograph (
http://www.mcponline.org/content/11/11/1468.long).
The problem comes when you are isolating only part of the isotope distribution. Then the isotope distribution on the product ions is not straight forward to predict. Because of this our first effort to do this will almost certainly be on the MS1 signal. Nick has been thinking of ways to combine the deisotoping with a fit in the time dimension too so that we can be better at handling incompletely resolved extracted ion chromatograms.
Brendan is right that I am generally not a big fan of decreasing selectivity as part of the routine targeted analysis. My preference has often been to just use product ions that don't have interference as opposed to adding a much more complicated signal processing step that would add variance (noise) to the measurement. Increasing the ion counts often increases the precision as long as it doesn't come at the expense of chemical interference. Additionally on trapping instruments the AGC reduces the fill time and it can let you scan faster without sacrificing sensitivity. That said, we are becoming big adopters of DIA workflows and the line between DIA and PRM is becoming blurred with methods like MSX (
http://www.nature.com/nmeth/journal/v10/n8/full/nmeth.2528.html?foxtrotcallback=true). So we are considering all of these classic signal processing methods.
Thanks for the post. We don't have a solution that can be used for deisotoping within Skyline at the moment. That said, it is something that is very much on our mind.
-Mike