My vote: yes, interference on y4 (and probably y6 also).
The other transitions are relatively consistent, but y4 goes from similar in abundance to y5 or essentially the least abundant to similar y8 or nearly the most abundant or a 2-4 fold change in relative peak area.
Also, worth noting that y4 as the shortest ion is the most likely to see interference.
Then I also noticed that y5 and y6 flip in their relative abundance.
Are there any other transitions you can detect for this peptide? If not, you might try a triggered MS/MS experiment to try to get a full spectrum on what is being measured, which would be expected to be chimeric, if this is interference. But, you might be able to learn more from a spectrum like that, especially if it turns out to have many large peaks that are not explained by your peptide.
Those are my thoughts. Maybe someone else will chime in with better ideas. Thanks for presenting your data on the Skyline support board.
--Brendan |