Why can't the predicted retention time be displayed? caixue  2023-12-25 20:48
 

hello,

I created a CiRT predictor with an r value of 0.99, but the predicted RT can not display as shown in attachments.

I don't know if there is a problem with the parameters or settings.

Looking forward to response

Xue Cai

 
 
Nick Shulman responded:  2023-12-26 04:26
Can you send us your Skyline document?
In Skyline you can use the menu item:
File > Share
to create a .zip file containing your Skyline document and supporting files including extracted chromatograms, iRT databases and spectral libraries.

If that .zip file is less than 50MB you can attach it to this support request.
You can upload larger files here:
https://skyline.ms/files.url

After we see your Skyline document we will probably be able to figure out what is going on.
-- Nick
 
caixue responded:  2023-12-26 05:25
Thank you for your response.

I have uploaded the zip file named FAIMS_TPD_cirt_prediction.sky in the URL: https://skyline.ms/files.url
 
Nick Shulman responded:  2023-12-26 08:22
Thank you for sending that Skyline document.

If you go to:
View > Retention Times > Score to Run
it is possible to see the iRT scores of your peptides.

When the "Retention Times - Score To Run Regression" window first comes up, by default, Skyline looks through all of the retention time predictors that it knows about and chooses the predictor which best fits the data. This is the reason that you are seeing "SSRCalc 3.0 (100A)" on the X-axis.
However, you can tell Skyline to use the iRT database that you have by choosing "Calculator > FAIMS_TPD_lib-assay" from the right-click menu.

When you are using the "FAIMS_TPD_lib-assay" predictor in the "Retention Times - Score To Run Regression", Skyline will display on the graph some statistics about how good of a fit the linear regression value was. Skyline tells you that the "r" value (i.e. the Pearson correlation coefficient) of that linear regression was 0.9623.
Whenever the r value of the linear regression is less than 0.99, Skyline will try to exclude a few points in order to get an r value which is greater than 0.99.

If Skyline is unable to find a set of points which result in a Pearson correlation coefficient which is greater than 0.99, then Skyline will not display the predicted retention time on the chromatogram graph.

You can tell Skyline to change the iRT values for these peptides by going to:
Settings > Peptide Settings > Prediction
and then push the calculator button and "Edit Current"
and then push the "Recalibrate" button and then "Use results"

After you do that, and press OK on all of the settings dialogs, you will be able to see the predicted retention time on the chromatogram graph.
Also, the r value of the linear regression in the "Retention Times - Score To Run Regression" window will now be equal to 1.
-- Nick
 
caixue responded:  2023-12-26 18:44
Thank you for your patient response.

I have resolved this problem. Thanks again.