target per frame Antoine  2021-09-07 09:06
 

Hello Skyline team,

For prm-PASEF
May I ask if the targets per frame graph corresponds to the max, the average or the median number of targets per frame within an MS cycle?

Thanks

 
 
Nick Shulman responded:  2021-09-07 14:37
I am not sure what the answer to your question is, but if you send me a screenshot of what you are looking at, as well as your Skyline document, I can try to figure it out.

In Skyline you can use the menu item:
File > Share
to create a .zip file containing your Skyline document and supporting files including extracted chromatograms.

If that .zip file is less than 50MB you can attach it to this support request. Otherwise, you can upload it here:
https://skyline.ms

It would also be helpful if you could attach a screenshot to this support request since I am not sure exactly what you are looking at.
-- Nick
 
Antoine responded:  2021-09-08 00:34
screenshot attached
 
Kaipo Tamura responded:  2021-09-09 18:01
Hi Antoine,
This graph shows the average number of targets per frame.

Thanks,
Kaipo