EIC of product ions with multiple neutral loss lorrain  2018-04-11 10:55
 
Hi,

I encounter this problem in both PRM analysis and MRM method setup when i was trying to look for a product ion which can have more than one neutral loss (for example, a y7 ion with two phosphorylation sites). The peptide and transition settings can predict the product ion with either one of the neutral loss (y7-98), but not both (y7-196). I was just wondering is there a way that we can have the skyline do that, or we can manually put in a product ion with both neutral loss (-196) in the list (especially for PRM analysis).

Thanks very much in advance!

Lorrain
 
 
Nick Shulman responded:  2018-04-11 21:25
On the settings page:
Settings > Peptide Settings > Modifications
there's a textbox "Max neutral losses".

If you set that to a number higher than the default "1", you should see the losses that you are looking for.