Hi Jung,
That is not great RT reproducibility. If those runs were don consecutively, then you should definitely look at what you are doing for chromatography. Perhaps your column is not fully conditioned for your sample on the first run?
In the field, <20% CV, especially for just 3 replicates, is considered acceptable, but your median across many peptides should definitely be considerably lower. So 30% is not good.
I would probably try running more replicates. Look at your chromatography for sure.
In the one set of chromatograms you show, it appears to me that the transitions you have chosen are experiencing interference on either side of your target peptide. Look at the shoulders (bumps) on either side that occur on only some of the transitions.
So, it seems there may be multiple reasons for your poor reproducibility. It should be possible to get much higher reproducibility out of PRM on a Q Exactive instrument.
Good luck troubleshooting your issues. Thanks for posting to the Skyline support board.