Hey Brian,
I finally got around to come up with some examples (as screenshots and partly with raw data) to demonstrate our problems with low concentration/intensity samples and the difference in detection of Skyline and Xcalibur/TraceFinder. I attached screenshots and corresponding raw data. However, unfortunately I don't have clearance to share data of the second example.
Example 1 ("uridine-neg_K01" and "uridine-neg_K03") suffers from RT drifts (as generally observable in the low concentration samples of the corresponding data set). Using Thermo's TraceFinder (basically Xcalibur algorithms), I'm able to detect the correct peaks by detecting the highest peak instead of the nearest and in case of K01 by additionally increasing the smoothing.
Example 2 ("uridine_CalG-04") features less drift, however the peak is really jagged and thus only a part is selected by Skyline. In contrast, selecting the highest peak in TraceFinder would provide me at least a proper peak height. Alternatively, increasing the smoothing even results in the whole peal detected.
I think having the options for smoothing and selecting the highest peak will result in highly improved automated peak detection. And since higher concentration/intensity feature clearer chromatograms anyway, optimizing low intensity peak detection with these parameters won't affect them. Except in special cases like isoforms maybe, but in the end an option is called like that for a reason ;-).
Other than that I really prefer Skyline over Xcalibur/TraceFinder due to several reasons. However, this issue in particular is holding me back in using Skyline regularly and eliminates my leverage to establish it in my lab. So I hope my examples provide proper insight into the issue and demonstrate the usefulness of these parameters :-). Please tell me, if you need any more information.
Best, Mathias
Little side note: all your mails (@proteinms.net) are considered as spam in Gmail. It complains that your mails are not authenticated (
https://support.google.com/mail/answer/180707?visit_id=1-636548254540712293-22150597&p=email_auth&hl=en&rd=1).