CCS calculations ? blaine roberts  2017-08-17 16:32
 
Is there a way to convert the drift times from ion mobility spectra to calculated CCS within skyline?
 
 
Brian Pratt responded:  2017-08-17 16:46
It depends on your equipment. We rely on the instrument calibration for this conversion, and at this point only Agilent provides that functionality in their API.

Where are you obtaining the drift information from?

Thanks for using the Skyline support board!

Brian Pratt
 
blaine roberts responded:  2017-08-17 17:42
I am using the Agilent 6560 and have the calibration files required. So I have the tfix and beta values that are appropriate for my ion mobility data.
 
blaine roberts responded:  2017-08-17 17:44
Was hoping to be able to convert the drift times to CCS within skyline for several reasons but the first was to build a peptide library that includes CCS.
 
Brian Pratt responded:  2017-08-18 05:08
I may need to make a small tweak to Skyline, to make sure the provided drift values are converted to CCS during Skyline chromatogram extraction, but otherwise we should be in good shape. May I see your various files to be sure it's all in working order?

Thanks,

Brian
 
blaine roberts responded:  2017-08-20 16:25
How shall I send you the files?

I plan on including the skyline file, calibration and data file. do you need anything else?

Blaine