Kaipo Tamura responded: |
2021-07-29 10:52 |
Hi Chi,
A single optimization library can contain optimizations values of different types. From your screenshot, if you change the optimization type to Compensation Voltage and add the CV values, the optdb file should end up containing both CE and CV values.
Thanks,
Kaipo
|
|
chi nguyen responded: |
2021-08-04 02:48 |
Hi Kaipo,
Thank you for your advice. It works fine for me to have both CV and CE information in one optimization library.
However, as I tried to perform PRM on Eclipse with FAIMS device. I can not export the isolation list with FAIMS information. I chose Thermo Fusion as instrument (see screenshot). There, I have no option "write Faims value" to choose. Do you know how I can do this? For now I added faims values externally, not in skyline.
Thank you
Chi
|
|
|
Kaipo Tamura responded: |
2021-08-04 15:07 |
Hi Chi,
I guess we haven't added this option for the Thermo Fusion yet. Do you know what the column header should be (e.g. for Thermo Quantiva it is "FAIMS CV (V)") or do you have an example isolation list containing the FAIMS values?
Thanks,
Kaipo
|
|
chi nguyen responded: |
2021-08-05 00:47 |
Hi Kaipo,
I attached a table with FAIMS information exported from the instrument method. I used eclipse with FAIMS.
regards
Chi
|
|
|
Kaipo Tamura responded: |
2021-08-05 10:27 |
Hi Chi,
Thank you for the example. I will add this option for the Thermo Fusion.
Thanks,
Kaipo
|
|
|
|